My HP 1663A logic Analyzer does not start up after the flash ROM update. The power up self test runs okay and after that a grey screen with some strange signs.
Is there a way to re-run the flash ROM update at startup?
« Last Edit: February 03, 2021, 08:08:38 pm by imac »
Re: HP 1663A logic Analyzer not starting up after flash ROM update.
« Reply #1 on: February 03, 2021, 08:02:34 pm »
The update was okay.
« Reply #2 on: September 28, 2021, 06:28:59 pm »
Is there anyone that have a dump of the eprom U14 01660-80004?
Offline fzabkar
« Reply #3 on: September 28, 2021, 08:53:12 pm »
Could we see your flash dump and update files?
« Last Edit: September 30, 2021, 05:07:38 am by imac »
Offline fzabkar
« Reply #5 on: September 30, 2021, 08:15:07 pm »
I can see how the files are structured. There is a 0x200-byte header after which is the actual payload. The payload is delivered in 0x100-byte blocks, with the first big-endian word being the byte count for that block. All blocks except the last have a byte count of 0xFE (= 0x100 - two bytes for the count).
If you were to provide a flash dump, I could verify whether the payload has been unpacked correctly. What surprises me is that there doesn't appear to be any checksum.
I have attached the unpacked files, as I see them.
« Last Edit: October 01, 2021, 09:57:34 pm by fzabkar »