So in the past years I went through several other digital oscilloscopes with higher bandwidth. Scopes come scopes go but the 1720A remains the same, I guess we grown together...
That said, this problem is still not solved ultimately.
A few addition, it turned out that the self rise time of my pulse generator is ~400ps and considering the scope should have 1.3ns rise time with this setup I should be able to see 1.4 ns rise time on the screen, so this is the goal.
In 2016 I bought a complete vertical amplifier board for this scope, and verified that isn't the problem.
Tggzzz's suggestion looks to be true indeed as poking around the delay line seems to solve the problem, but only temporarily...
But then look into this problem deeper, how could the delay line affect the bandwidth? When I got the replacement vertical amp board it came with a piece of the delay line cut, so I could examine it a little.
In this scope the delay line is constucted to have 165 Ohm impedance according to the manual. it is a differential coax (twinax?) cable. The dielectric material is rather hard plastic with opaque color, the inner inductors are solid, by the look of it silver plated copper. The shield is a generic tin plated copper braid.
Now what is exactly the role of the shielding in terms of bandwidth?
- Since the drive is differential it should not carry useful signal.
- As I see its purpose mainly to define the impedance of the cable, but then if the impedance is not matched the rise time is not affected (as I experienced so far). Anyway from this point of view it makes no difference whether the braid inductors are isolated from eachother or not, so long they are connected to ground at either end, right?