So my HP 1740A is displaying the voltage exactly 10x too small.
If the volts per div is set to .1 and the probe is connected to the 1V Cal, I get a one division reading (same for both A & B).
The amplitude of the reading is exactly 10x smaller than it should be and I can't seem to correct it.
Help
What I know:
Probe is a Tek P6122 (10x)
1V Cal has been set to .500 volts with a Fluke 177 DMM
Depressing the blue 10x button on the scope increases the width of the signal.
Depressing the tan 5x button increases the amplitude by 5x.