Hi,
I'm trying to fix a currently non-working HP 3325A (rev C). I have the manual (unfortunately only available for rev F - but the design changes are not huge in the areas I have been working). I started as per the trouble shooting procedure and I have isolated the power supply board from all other boards and have established that there is a problem there.
The +15/-15V and their sense rails appear to be working fine. Stable voltage within 0.1V of +/-15V
The 5V rail and 5V sense rail are however sitting at 11.49V which is the same as the unregulated, rectified supply from the transformer winding. I've been over just about every component in the 5V regulator circuit and can't find any problems. I've replaced most of the components that could be faulty including the main pass transistor (Q1), the op-amps that regulate it, the filter caps and the SCR/zeners. Nothing I have done seems to bring the 5V rail back into regulation. The 5V supply is referenced from the -15V rail but given this is OK I don't see that the problem would be there - I have checked the +/-15V circuit main components (cap, transistors) and it all seems to be fine.
It's driving me a bit potty! Can anybody give me any pointers or pearls of wisdom please?
many thanks
Tony