Greetings all,
I have a HP 34401a that fails 612, 613, 615, 617 and 619. It also does not respond to DCV or Ohms applied voltages/resistance measurements. ACV appears to work as expected. I have not yet tried amps.
What I have done so far:
I searched this forum for all references to the above mentioned codes, and while very informative, did not quite match or solve this particular problem. I checked the supply voltages, voltage across 2W Hi/Lo in the different Ohms ranges, and all voltages appear within range.
Studying the schematics and block diagrams shown in the service manual, I thought it odd that the unit would display/read ACV but not DCV, as the ADC receives only DCV (i.e., RMS ACV are converted to DCV per the manual). So I decided to do the following:
Applied approx. +5VDC from my benchtop adjustable regulated power supply, and (using a handheld DMM or 'HHDMM') measured between the large PCB ground pad located near CR202 and R104 (on the end that connects to K103).
With the 34401a in DCV mode, it displayed correctly, but only after I touched the probe tip of the HHDMM. The HHDMM also confirmed that the applied volts were present at R104. I also could hear what sounded like one of the relays 'click' at the same time the 34401a correctly displayed the applied voltage (hope all that makes sense...).
So I'm wondering if this is a logic problem, bad relay, or ...? After writing all this, it looks as though the relay coils are supplied from separate +- 15VDC supply? I will be attempting to investigate that next.
The error codes are perplexing, as it appears the 34401a is unable to read the Ohms compliance voltage (something Dr Frank mentions in a related post...). This caused me to suspect the DC amp, but now in light of the above, I'm uncertain what's causing them.
Any suggestions/advice on what to check, would be greatly appreciated... I think I might be getting close to zeroing in on the culprit, but not quite certain where to look next.
Thanks.
Vern