HP 34401a Self Test Fail
lkmurdoch:
Hi everyone,
My first post so here goes.....
I have a HP 34401a that fails self test. It does measure AC volts, DC volts reasonably accurately (very close to a Fluke 45) but Ohms is way off cal.
The Ohms function measure 00.000 when the leads are shorted, but nothing else. I have checked for signs of obvious damage internally and checked the
power supply voltages which are fine. Any suggestions gentlemen?
Thanks in advance,
Mike
free_electron:
error code ?
Dr. Frank:
Hi and welcome..
not possible to figure out. I suggest you give more detailed hints:
- do you run the self test, and then you get an error?
- error code available?
- what does the instrument read in the different Ohm ranges, each 2W and 4W, with shorted leads and with resistors of appropriate value attached?
- did you check the same on the rear connectors?
Frank
tszaboo:
Download service manual from Agilent. Check for your error code in that.
lkmurdoch:
Hello all,
No error code was displayed after the self test.
I checked the Ohms current source opamp (U201) for correct supply and reference voltages and they are correct.
Interestingly on a whim I checked for a voltage on the test leads of any kind and the is none.....the Fluke 45 on the bench shows 1.56 volts DC.
Therefore I suspect that there is indeed an issue with the Ohms current source.
I would appreciate any insights any of you my have to offer.
Michael