Author Topic: Protek 506 resistance measurement confusion  (Read 639 times)
Offline W1BCCTopic starter
« on: February 26, 2021, 02:55:09 pm »
Protek 506 - used, but new to me.  Nice rubber case and leads.
My Fluke 175 says a 10k resistor is 10k.
Using same leads, the 506 says 11.8k.  Shorting = 1.8k
Applied a thorough squint to the manual.  No calibration or reset options noted.  Even swapped out the fuses,
What happened?
Thanks,
--Brad W1BCC
The following users thanked this post: Kartika
Offline retiredcaps
Re: Protek 506 resistance measurement confusion
« Reply #1 on: February 26, 2021, 08:36:41 pm »
I suggest checking the input protection starting with the MOVs and PTCs.  A good MOV should measure 0L ohm using Fluke 175.  A good PTC should measure between 500 ohm and 1500 ohm (an estimate on my part).  Both can be done in circuit.
Offline srb1954
  • Retired Electronics Design Engineer
« Reply #2 on: February 26, 2021, 08:51:28 pm »
What happened?
Thanks,
--Brad W1BCC
Does the meter give correct readings on the volts and current ranges?
Check the voltage drop across the meter when measuring current. Does it comply with the burden voltage specified for the meter?
Try a different set of test leads.
Check and clean the input jacks.
Re: Protek 506 resistance measurement confusion - Update
« Reply #3 on: March 03, 2021, 04:02:58 pm »
Yes, tried swapping leads.  Also a little DeoxIt.
Appears to measures resistance accurately, just with an additional 1.7k.
Otherwise, it seems to function properly.
Thanks for all the suggestions,
--Brad  W1BCC