Hello,
the first error relates to dead batteries inside the nvSRAMs (DALLAS).
This has been discussed in several forums, how to save the content of the cal-RAM, and reprogram it into new parts.
As these batteries last between 10 and 20 years, a 1989 instrument should have been repaired long time ago.
The second error is not common; it's the first time, that I heard of that.
Most probably, it's located on the A5 outguard controller (main processor), as display and hardware are not controlled correctly.
Well possible, that one EPROM lost its content.
Please visit xdevs site, he also hosts all FW versions:
https://xdevs.com/fix/hp3458a/For a first attempt, you may update your six EPROMs inside your unit to the latest FW 8.
FW 9 is also available, but that's not necessary.
Frank