The last of the trilogy of cheap equipment I picked up at the hamfest last Saturday. While waiting for parts on some other stuff, I decided to dive into this. It's a modular multimeter, with one of the old dot-LED displays and a choice of several plug-on modules such as a DCV unit, an AC/DC/Ω unit, and a BCD output (before the days of HPIB). This one had the standard multimeter module and the 5-digit display; for $10 I couldn't pass it up.
Quick testing shows DCV and ACV to be accurate on all ranges. I had some flaky readings at first, so I hit the switches with Deoxit and that seemed to tame them. Ohms is good on the 200Ω and 1KΩ ranges (short and open behave as expected), then gives nonsense readings on the higher steps. The PSU gives good ±5 and ±12 values, so I dived into the troubleshooting tree. Seems like the voltages around Q15 are a bit off, though only about 10-20%. The guide says U2, Q15, Q16, and CR5 - but looking at the schematic, CR5 is in the ACV section, so I think they really mean CR15. They're all common parts except for Q15 which is the HP 1855-0308 dual FET found in several meters.
So I can probably scrounge up the other stuff, but this one seems to be a problem - there's no readily known substitution for it that I can find. Used parts are fairly readily available around $15 or so, but I'm not too excited about sticking a used part of unknown quality in something I'm trying to troubleshoot. Also, I hate to make multiple orders for a repair just to find out the fault wasn't where I thought it was; I've done that enough as it is.
Does anyone have a suggestion about an appropriate replacement part for this? It's a dual matched depletion mode JFET, probably 30-40V rated.