Author Topic: HP 3478A / Calibration failure  (Read 1964 times)

0 Members and 1 Guest are viewing this topic.

Offline Eltax1693Topic starter

  • Contributor
  • Posts: 22
  • Country: se
HP 3478A / Calibration failure
« on: September 21, 2016, 08:17:17 am »
I bought my HP 3478A from Ebay a month ago. It passes the self test, so I sent it to a cal lab. [replaced the battery too]
Everything went fine, except the ACV and ACA ranges. "not possible to adjust".

What can be wrong?

TNX
 

Offline singapol

  • Frequent Contributor
  • **
  • Posts: 420
  • Country: sg
Re: HP 3478A / Calibration failure
« Reply #1 on: September 21, 2016, 10:05:00 am »
I bought my HP 3478A from Ebay a month ago. It passes the self test, so I sent it to a cal lab. [replaced the battery too]
Everything went fine, except the ACV and ACA ranges. "not possible to adjust".

What can be wrong?

TNX

The self test only test display , memory and A/D convertor communication only. Read chapter 3-60 page 3-7
of the service manual on self test.. Why Ac volt and current cannot calibrate? Unfortunately  I think AC measurement circuit has  been damage by previous owner overloading input. Read Service group B - AC volts and Ac current troubleshooting on page 7b-1 of sevice manual. Hope you can find a good technician to repair HP3478A. :(

HP 3478A service manual-
http://www.arimi.it/wp-content/Strumenti/HP/Multimetri/hp-3478a-Service.pdf
 

Offline dom0

  • Super Contributor
  • ***
  • Posts: 1483
  • Country: 00
Re: HP 3478A / Calibration failure
« Reply #2 on: September 21, 2016, 05:58:15 pm »
iirc the self test does show whether valid calibration data was loaded from the cal SRAM.
,
 

Offline Eltax1693Topic starter

  • Contributor
  • Posts: 22
  • Country: se
Re: HP 3478A / Calibration failure
« Reply #3 on: September 24, 2016, 11:36:49 am »
OK, tnx for your help.
 