I have my HP3561a on the bench because of a Test error and just wondered if anyone else has experienced this issue.
When I power on the unit the self test returns Error # 07  28   Ram: G
In the Service Manual this comes up as Display memory error in NMOS 16K dynamic ram module, G position.
Is this the only module I should be concerned with or will there be more problems,
« Last Edit: April 07, 2016, 01:51:30 pm by Johnny10 »
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
Online PA0PBZ
Re: HP 3561a Signal Analyzer Display Troubleshooting
« Reply #1 on: March 30, 2016, 01:25:15 pm »
Looks like a classic example of video ram problems, it could be a stuck address/data line but most probably just the ram chip.
« Reply #2 on: April 07, 2016, 01:50:47 pm »
OK, Solved Post.
I exchanged Ram board with one from another of my 3561A units.
Everything back to normal on the display.
Now I have another question which I will post in a new thread.
Thanks
« Last Edit: April 07, 2016, 02:16:58 pm by Johnny10 »