I bought a HP 410C from the sixties in fairly good condition.
DC V and A and Ohms are very good but I am not able to calibrate the AC V ranges.
The supply voltages are all inside their specified range.
The first issue is that I am hardly able to do the AC zero alignment, the coarse adjustment pot A3R31is almost at the cw limit, I had to bypass the A1R7 resistor to have any headroom for adjustment.
The 0.5V, 1.5V, 5V and 15+V trimmers are at their ccw limit to just or not quite get a correct reading.
Is anyone here familiar with this specific piece of equipment?
First I suspected the AC probe to have somehow deteriorated, but even when feeding DC voltage into the AC socket, I am not able to do an alignment.
Furthermore I noticed that the Orange Drop capacitor A3C1 is quite sensitive to temperature changes - do those fail this way?
Any help would be appreciated.