Hey there fellow techies, first post!
I have recently bought an old HP 4194A, upon arrival and inspection i found it gives two error codes, one is bridge unbalanced, other is AC current overload. my guess being that some klutz has blown up the front-end of the impedance analyzer.
As many of you will know this aint such an easy device to fix, so i'm asking on the forum if anyone has repaired this issue before, and what a possible remedy could be. today i have spend the better part of the day to go over the service manual and re-adjusting the voltage / signal levels over the unknown ports and internally over the test pins. here is what i found:
1- PSU seems intact, all voltages well within spec
2- CRT works, be it a bit dimm
3- device starts up fine, logic boards seem in order for as far as i can tell.
4- adjustments in the manual on fine-tuning the bridge balance have been completed, several times, with success, i can adjust A23 board TP 15 TP 16 to the desired voltages / signals and have the correct signal on Lcurr port.
only thing that is left is to check the sensing circuitry, could be a faulty op-amp maybe? or a protection diode? i need to find schematic first as it would help allot, and my spectrum analyzer is of the analog kind, and that doesn't help, so i cannot complete all tests, since i cannot really measure Decibels with my DOSX2024
any tips, tricks, etc are most welcome, this device will be used for medical research (skin / tissue impedance VS inflammation, etc. ) so once its fixed, it will benefit more people that just myself.