I found my HP 4275A failing Zero Cal with ERR1 on 10khz and erratic short value on almost all frequency.
Looking on the service manual, after some investigation i ended up on A6 Board (Oscillator), input of the filter(TP2) was OK, on TP1 and TP9 i found the same problem, 10khz amplitude way too high.
After further investigation i found both filter not working as expected because lack of amplitude level control, at this point i was a little bit confused, every thing on both filter seemed to work just fine and i started to look at the Photocell CP1 and CP2, as far i understood from the schematic and the block diagram discussion the amplitude level control is performed adjusting the frequency roll off through the variable resistor on the photocells CP1 and CP2, automatic level control work with the feedback feeding the led of the photocells, brilliant design !!!
Measuring the photocells i found CP1 resistance value high and variable around 200kohm, CP2 resistance value around 20kohm, photocells are marked MCD-521 YOKOGAWA brand, of course no way to find the datasheet, since is there no power on the resistor of the photocells i guess is this issue is just related to the aging.
Even if with strange value i got good response feeding the led, the low side of the resistance was OK (high side frequency roll off), at this point was clear to me that the problem was related to the too low resistance on the photocells with no light from the led (low side frequency roll off).
With complete understanding of the issue has been pretty easy to fix it working on the filter capacitor components C6, C8, C26 and C28, looking amplitude and settling time with the scope i ended up adding 0.02uf on C6-C26 and 1000pf on C8-C28, to further improve the high ranges i added 100pf from the photocells exit to ground.
Finally my 4275A pass all of the tests and the Zero Cal is perfect !!!
Of course i am still concerned, i consider my fix as temporary and i am looking to replace the photocells, it would be very useful if someone have some knowledge of the spec of this photocell type
I hope this repair could help someone!!