Author Topic: Fluke 8840a/af 4-wire resistance inoperative  (Read 2407 times)

0 Members and 1 Guest are viewing this topic.

Offline ExcavatoreeTopic starter

  • Frequent Contributor
  • **
  • Posts: 901
  • Country: us
Fluke 8840a/af 4-wire resistance inoperative
« on: July 16, 2017, 12:52:19 am »
I have an old (35 years) Fluke 8840a/af bench meter that I purchased from an E-bay Fluke seller. (funny that I'd know of any, ha ha) 

As far as I can check, everything works  (voltage (AC and DC)  and 2 wire resistance) except the 4 wire resistance mode.   I can't believe it's much, but I've not had any luck finding a service manual. 

Can anyone give me a direction in which to start?  A 1.5 ohm resistor (confirmed using Dave's "poor man's 4 wire technique of using two meters")  reads .5 ohms. 
 

Offline alm

  • Super Contributor
  • ***
  • Posts: 2881
  • Country: 00
Re: Fluke 8840a/af 4-wire resistance inoperative
« Reply #1 on: July 16, 2017, 01:27:24 am »
Did the current from the force terminals match the specification? It is not a particularly tight spec, but it should not be 66% low :P. What voltage do you measure across the sense terminals?

This manual has schematics. I think the /AF version was quite similar (some options default and maybe tighter specs?).

If the values read too low only in 4W mode, then that suggests to me that there is a low resistance path between sense low and sense high. Maybe one of the protection JFETs (Q301/302) is leaky?
« Last Edit: July 16, 2017, 01:30:26 am by alm »
 

Offline ExcavatoreeTopic starter

  • Frequent Contributor
  • **
  • Posts: 901
  • Country: us
Re: Fluke 8840a/af 4-wire resistance inoperative
« Reply #2 on: July 16, 2017, 05:23:16 pm »
Correction:  The 2 wire reading is 1.5xxx (correct) and the 4 wire reading is 1.2yyy.   The error does not scale.   Of course, it's so small for most measurements in the higher ranges, it appears that the 4 wire readings are correct.

I removed Q301, Q302 and clamp Q309, with no effect.

I checked Q303, and even substituted another part, but there was no effect.

I've got to be missing something.  I just can't figure out what could be causing the 4 wire reading to be wrong without affecting the 2 wire mode.
 

Offline alm

  • Super Contributor
  • ***
  • Posts: 2881
  • Country: 00
Re: Fluke 8840a/af 4-wire resistance inoperative
« Reply #3 on: July 16, 2017, 10:55:19 pm »
How do the voltage and current compare between 2w and 4w mode? I do not have the schematic in front of me, but what about any switching (e.g. mux) that is only used for 4w?

Offline RobK_NL

  • Frequent Contributor
  • **
  • Posts: 250
  • Country: nl
Re: Fluke 8840a/af 4-wire resistance inoperative
« Reply #4 on: July 17, 2017, 02:56:16 pm »
Have you checked the front/rear switch and U301D?
Tell us what problem you want to solve, not what solution you're having problems with
 

Offline ExcavatoreeTopic starter

  • Frequent Contributor
  • **
  • Posts: 901
  • Country: us
Re: Fluke 8840a/af 4-wire resistance inoperative
« Reply #5 on: July 18, 2017, 03:46:29 pm »
How do the voltage and current compare between 2w and 4w mode? I do not have the schematic in front of me, but what about any switching (e.g. mux) that is only used for 4w?

I can't see anything different so far.

Have you checked the front/rear switch and U301D?

I have bypassed the front/rear switch with no effect.
I'll be honest, I messed up (got turned around and or backwards) when checking voltages at U301 and noticed that when I connect pin 13 to ground via my 10 Megohm Fluke 87,  the reading is correct.  I'll be honest again and admit I have no idea what that could mean.

I've checked the Jfet switching between Q310 and Q301 via U305 and can't find any problems.



 

Offline alm

  • Super Contributor
  • ***
  • Posts: 2881
  • Country: 00
Re: Fluke 8840a/af 4-wire resistance inoperative
« Reply #6 on: July 18, 2017, 04:17:40 pm »
I can't see anything different so far.
So current source switching is fine, and the problem is somewhere at or downstream of S301.

I have bypassed the front/rear switch with no effect.
I'll be honest, I messed up (got turned around and or backwards) when checking voltages at U301 and noticed that when I connect pin 13 to ground via my 10 Megohm Fluke 87,  the reading is correct.  I'll be honest again and admit I have no idea what that could mean.
Messing up is an excellent way of finding a way out if you are stuck. Too bad it is often associated with extra work and expenses :(. Pin 13 is one of the input nodes to the ADC. Does it do the same if you ground pin 17 or 18? If not, then it sounds like a bad solder connection or break in trace/via.

According to the table on page 93, U301C and U301D are involved in switching between two and four-wire mode. An open U301D or shorted U301C might explain the symptoms. If grounding the S/H input fixes stuff, then my money is on an open/high-resistance U301D (pin 12/13), either a broken IC or connection. In 4W mode, you should measure close to 0 V between R314 (C306 side) and Q314 pin 3.

Offline xwarp

  • Frequent Contributor
  • **
  • Posts: 367
  • Country: us
Re: Fluke 8840a/af 4-wire resistance inoperative
« Reply #7 on: July 19, 2017, 07:30:32 am »
If you find any bad chips, send me a message as I have a couple 8840 boards that I have salvaged parts from.
 