« Reply #8 on: January 26, 2019, 04:46:23 pm »
OK
I connected the new Relay.
And look at that nice stable output on the integrator.
Changed the usual M1
All electrolytics.
Have 15.01 and -15.01
6.42 and -6.45 Reference
5.10 on the 5 volt line
Only one problem left to manage...
the -15V line is not always 15V sometimes on turn on it is -13.5 +-
Repeated on-off cycling and I get my -15.01 back.
When it is -15.01 it is stable until I cycle power switch.
Same when it is 13.5
Only thing I can think it might be now is that tantalum I haven't changed C11.
Guess I should work on that after lunch.