Howdy folks,
I purchased a HP 5300B with a 5308A 75MHz frequency counter (non-op) with the intention of repairing it and giving it a home on my bench.
There is an audible whine coming from the power supply area of the 5300B, so I checked over the power rails, electrolytics and tantalums.
I did have to replace a few, but it made no difference.
So I found a PDF of the manual (available through Keysight), and stepped through their troubleshooting flowchart.
They even make mention of the whining, so it apparently is a somewhat-known issue.
These units have a blocking-oscillator based PSU, with some half-wave rectification to give +/- 17V, +/- 5V, and +3.5V.
In the flowchart one of the tests is to apply +6V across the filtering capacitor and probe the collector of the switching transistor.
When I performed this I found almost double the peak-to-peak amplitudes specified (see image below).
Pulling the transistor and testing it with a cheapo component tester didn't raise any red-flags, though.
I continued through the power supply checks in the manual, testing nearly every component mentioned out-of-circuit to no avail.
Either the trouble is more subtle than expected or I bought a lemon.
Does anyone have any ideas or experience with the 5300B measuring systems?
If there are any other details that would be helpful, let me know.
Manual via Keysight:
https://www.keysight.com/ca/en/assets/9018-05900/user-manuals/9018-05900.pdfAttached are:
- a picture of the circuit board with the power supply section outlined in red
- A capture of the blocking oscillator transistor collector voltage (5V/div, 0.1mS/div)
- PNG of the power supply schematic, again the full manual can be had through the Keysight website.
Best,
Proton