Author Topic: HP 53131A - dead power supply  (Read 1839 times)
Offline fchkTopic starter
« on: June 07, 2020, 09:56:36 am »
I've got an HP 53131A Universal Counter with a dead power supply. Does anybody has schematics for the power supply? The service Manual only has schematics for the mainboard. Are there common failures known?
There are two fuses, but both are ok. Capacitors look ok at first glance, but a a whole re-cap would be my first guess.
Can I run the power supply without load?
Any recommendations?
fchk
Offline George Edmonds
Re: HP 53131A - dead power supply
« Reply #1 on: June 07, 2020, 11:48:11 am »
Hi
The power supply is a third party item and there are no power supply schematics in the CLIP.
Be very careful as it is always on to power the OCXO with a soft power switch for the counter itself.
It is of the "flyback" type and will readily self destruct, it may need a small minimum load.
George G6HIG
Offline Qw3rtzuiop
« Reply #2 on: June 07, 2020, 01:22:38 pm »
Offline DeepLink
« Reply #3 on: June 07, 2020, 07:55:54 pm »
Use some resistors as load ~10% of rated load is sufficient to have proper regulation
Look at the start up resistor R2, can be open
And do exchange all the caps (mine was leaking badly - even though the PSU was working)
Use proper SMPS caps - low esr, high ripple current
As the PSU is always on (and the fan is always running), it is possible to make a mod that shut down the PSU/Fan
The fan can be upgrade to a low noise, as the original is quite loud
Offline Decee1
« Reply #4 on: March 04, 2022, 04:36:11 pm »
Hello,
I've just finished repairing my power supply and i thought i would leave some information about the fault.
Fault description:
No voltage on secoundary side, cannot measure any pulses on IC1 pin 6(out) but IC1 pin 4(RT/CT) is available and running with sawtooth wave.
Fault conclusion:
Q22 is acting like a short but is still fully functional. In my case it measured 1V on collector and emitter and 0.5V on base.
Measurements:
if pulses is not on IC1's output, RT/CT is active and there are no voltage on comp pin IC1 does as expected and therefor Q24 pulls IC1 pin 1 (comp) down to GND, then measure base of Q23 or R20, then it is measuring around 0.7V or above which turns on Q23, If Q22 measures the same voltage across it, the Q22 is faulty and should be replaced
Fault explained:
Q23 is 'on' which causes Q24 to 'turn on', pulls down the internal current source of IC1 pin 1(comp) and therefore 0V on IC1 pin 1(COMP) pin and no switching pulses on IC1 pin 6(out)
Hope my debugging can be useful for somebody else.
Stefan
The following users thanked this post: gamalot, alm