Recently, I got a HP 54503A with a known NVRAM failure (empty battery).
After replacement against a new NVRAM the scope passed the self-test (see image 1).
After initialization of the self-calibration I encountered the following errors on ch#1, whereas ch#4, ch#2 and ch#3 pass the calibration (see images 2 to 5).
The message “Check BNC cable connection” has been verified and cable connection found OK (otherwise the calibration could not pass).
The self-calibration page shows the following (see image 6) status summary (the delay calibration was not executable without pass of all 4 channels).
The probe calibration menu (see image 7) displays an error: “Attenuation too high or bad connection on channel #1”.
The connection to the calibration output has been rechecked and found OK. Channel #1 shows the beam only, but no applied signal. Channel #4, #2 & #3 display signals applied.
I assume channel #1 is obviously dead as no applied signal (beam only) is displayed.
I disassembled (lifted from the main board) the channel #1 cover with the attached BNC connector and visually inspected the cover inside and the board area for sign of marks or blown components. There was no evidence of any visible damage.
Any repair hints or ideas how to fix this problem?
Does someone has access to the drawing/schematics of the input circuit for troubleshooting or is familiar with this type of failure?
Where to get affordable surplus parts?
Any suggestions are highly appreciated.