I finally got to looking at a for-parts HP 6236A I picked up and not-surprisingly the thing still mostly works - with the exception of the -20v supply (The 6 and +20v supplies work great... voltage is adjustable and when connected to a load they supply the rated amperage, 2.5A/0.5A)
The -20 supply is a constant -31 volts, and I cannot adjust it... when -20 is selected on the meter, it pegs to full on. I don't run it for long because I noticed Q4 (driver) gets pretty hot (this has got to be a clue). I followed the troubleshooting steps in the service manual - without skipping any! - and I'm at a dead-end.
Steps in 5-3 (initial troubleshooting) directed me to 5-6 (-20 troubleshooting).
section 5-6: High Output Voltage
step 1: shorting Q13 brings the -20v supply to 0v (and the -20v meter drops to zero as well) so it says to continue to step 2 (i later came back to this and removed and checked Q3, Q13 and CR29 anyway - all OK)
step 2: TP5 is -0.141, so (a) says check defective Q4.
Q4 tested OK in-circuit and out of circuit (I later replaced it with one from another supply - no change). That seems to be the end of the troubleshooting, but I continued to step 3
step 3: TP14 is -0.7, so (a) says check for open CR26 or bad U2.
CR26 tested fine, as did U2. Someone socketed the op-amps at some point

so I took out U2, built up a test circuit on a breadboard and verified with my scope that it works as a comparator and an amplifier. The amp was a bit noisy so I swapped it with U4 but that made no difference and the 6v supply still works great. (b) says check Z1, which I also did just for kicks.
I then went through section 5-4 to check the bias circuitry and all the voltages are in-spec, i got TP1=7.19, TP2=-6.22, TP3=-12.40. *TP1 is low, but still within 5%.
I "think" that was the end of the troubleshooting in the manual, so I did some more testing on my own...
- i see that the transformer is putting out 25v volts at the center tap (pin13/15 to pin14) and 50v across 13/15
- i checked every CRxx I could find (in-circuit) and they were OK.
- also pulled out a leg of CR26 and it was OK
- pulled out a leg of C12 and it tested fine at 7uF
Looking at the schematic, I can't see anything else that would cause this -31v behavior. I know the big caps are often a problem with this old equipment, so I tested C17 and C7 from the +20 supply and they were close to each other - 1712uF and 1585uF respectively (esr on my junk tester showed .18/.16)
I'm at a loss, can I remove any/all of C11, C12, C13, C14, C17, C18, or C19 and run the supply without doing more damage to anything?
Any other ideas?
Thanks for any advice!
-Tom