Poll
Any ideas on how i could possibly fix my hp 6632a system dc power supply that has a test error 18?
Send it to a repair specialist
1 (100%)
Chuck it in the trash/ no good!
0 (0%)
Total Members Voted: 1
Author Topic: Hp 6632a error test 18  (Read 1989 times)
Offline Jesset1996.5Topic starter
« on: July 08, 2015, 03:53:42 am »
Hello everyone.  I have a hp 6632a that has a test error 18 message on the lcd once it is turned on. I have already tried simply clearing the error with no luck. Any suggestions on what i should do to fix this?
Offline MarkL
Re: Hp 6632a error test 18
« Reply #1 on: July 08, 2015, 02:17:49 pm »
Download the service manual and follow the troubleshooting section in chapter 4:
  http://literature.cdn.keysight.com/litweb/pdf/5957-6365.pdf
The manual lists error 18 as:
  CV/CC zero too high.  CV or CC DAC probably defective (see Figure 4-11)
It doesn't sound promising, but you won't know for sure until you work through the troubleshooting flowchart in 4-11.
« Reply #2 on: July 08, 2015, 07:18:38 pm »
Thanks! I actually got it to work by using the bypass jumpers tp1 and tp2. Now i must figure out how to erase the previous programming that caused the problem  I actually just got this thing at a ham swap so its previous owner must have incorrectly programmed it!