i spended an other few hours on my display problem again. First of all, the following post i found on Keysight forum, keeps me away from the thought my CRT is the problem.
Q:
I have an 8563A that powers up OK but the display stays blank. It shows a green flash when powering down indicating a normal field collapse.
I have checked all the DC power supplies and also the signal levels on J201 202. Also the ext video out produces some output on a scope.
HV module substituted and A17 driver with no effect.
I suspect it is the CRT, is there any way to confirm this?
A: If you saw the green flash when turning the power off then your CRT is probably OK. I'd suspect the A2 controller (especially since your HV and driver assemblies seem alright)----------------------------
I also see the green bright flash/spot in my CRT when powering off so that answer sound reasonable enough for me to continue and when i poke arround U210 with my probe i see my text boxt on the right and bottom side light up as they should. So there is hope i would say...
In the following weeks, i decided to take out U207,U210 & U211 and put them in (high quality) round-pin IC-sockets for easy probing. I also took out the R250(3M16) resistor and R266(1M47) but that checked ok.
Shorting the switch U207 pins 2 and 3 my picture looks ok. So i decided to pull pin 1 out of the socket and traced down the switching line from 74HCT377-> DEF1.
the line was high where i suspected a low to enable the U207 switch.
Meanwhile some part came in which i ordered over a month ago. A Harris HI-> U207 and two MC34084B -> U210 and U211. I did a quick swap but without any succes.
i checked my +15VF supply voltages, but they are just +14.92 and -14.91. quit on the low site if you ask me. Normally i would suspect a few .1 over the printed values in the Manual but these are below.. not much but.
And my DEF1 line is high which open the U207 1,2,3 switch. strange..
suggestions?
i am also curious what exactly to be expected on the line (-U207B/P7- U207A/P3 - U210D/P13 see marked in picture.)