Author Topic: [SOLVED] HP 8563A Spectrum Analyzer Repair - Is the YIG oscillator dead?  (Read 7627 times)
Offline 1GhzTopic starter
  • Country: kr
EDIT: June 29, 2020
TL;DR
The problem was caused by bad YIG oscillator.
Replaced the YTO (YIG oscillator) and that fixed the problem.
I recently bought an used HP 8563A spectrum analyzer.
But it's not working and it displays an error messages.
301 YTO UNLK
317 FREQ ACC
334 LO AMPL
Here is the pictures.
1. 8563A 1st LO Output (picture from the 8563E)
1002784-0
It's not working correctly. No signal output from low to mid frequency range.
2. 8563A CAL Output (picture from the 8563E)
1002788-1
It's working correctly.
3. 8563A Power Supply LEDs
1002786-2
I think it's working correctly.
Is the YIG oscillator dead? Should I replace the YIG oscillator?
Thanks,
« Last Edit: June 29, 2020, 11:26:41 am by 1Ghz »
Offline 4thDoctorWhoFan
Re: HP 8563A Spectrum Analyzer Repair - Is the YIG oscillator dead?
« Reply #1 on: June 13, 2020, 06:19:56 pm »
Since you are displaying the cal signal correctly, it's not the YIG.  Usually with a YIG problem, you will not get any signals to display.
I'm no expert though.  :-//
Offline MadTux
« Reply #2 on: June 13, 2020, 08:42:26 pm »
Better look at block diagram of 8563A for YIG frequency sweep range before declaring YIG for dead.
YIG of 8561E runs between 3-6.81GHz, somewhat close to those 5-7GHz you measured, probably also depends on what band your SA is set to.
Low band 0-2.9GHz mixes it to 3.91GHz, so YIG should sweep beween 3.91-6.81GHz.
High band 2.9 GHz and above downconverts it to 310MHz, so it should sweep 3.21-6.11GHz
Harmonic mixing in bands above, probably similar to HP 8566B
http://www.simonsdialogs.com/2018/11/hp-8561e-spectrum-analyzer-30-hz-to-6-5-ghz-some-consequences-of-a-past-disaster/
http://www.simonsdialogs.com/wp-content/uploads/2018/11/8561e-block-diagram-with-annotations.jpg
Perhaps try setting the 8563A into zero span mode, then you can measure frequencies without sweep.
YIG coil can also be tuned by current from power supply for basic testing outside of instrument, if you get crazy otherwise ;-D
Your YIG is probably fine, btw, more likely the PLL is messed up, perhaps dried up capacitors or so, usually the first that goes bad if noone has messed anything up.
« Last Edit: June 13, 2020, 08:51:14 pm by MadTux »
  • Country: kr
« Reply #3 on: June 14, 2020, 10:02:42 am »
Normally, YIG of 8563A runs between 3 to 6.81 GHz, too.
More pictures.
1. 8563A low band 0 Hz to 2.9 GHz sweep test
1) Setup
1003058-0
2) Result
1003056-1
No low to mid frequency sweep.
2. 8563A high band 2.9 GHz to 6.2 GHz sweep test
1) Setup
1003062-2
2) Result
1003060-3
3. 8563A CAL output 800 MHz span test
1) Picture from the 8563A
1003064-4
No 0 Hz LO feedhtrough displayed.
No CAL output signal displayed.
2) Picture from the 8563E
1003066-5
4. 8563A CAL output zero span test
1003068-6
1003070-7
« Last Edit: June 14, 2020, 11:06:32 am by 1Ghz »
Offline MadTux
« Reply #4 on: June 14, 2020, 11:30:45 am »
PLL goes crazy on that one, YIG likely is fine.
Try measuring tuning current (as voltage) on YIG tuning coil without sweeping, it should be a constant current(voltage), on your instrument, it's likely oscillating like crazy.
The spikes you see are probably when unlocked PLL matches frequency at a sweep spot. A few MHz above or below on 8563E sweep, 8563A YIG has drifted to a totally different frequency, so it generates the wrong impression of no YIG output, when in reality it's just running all over the spectrum.
« Last Edit: June 14, 2020, 11:39:32 am by MadTux »
  • Country: kr
« Reply #5 on: June 29, 2020, 11:13:48 am »
I bought the YTO from Test Cal Instruments.
1. -5 dBm 20 GHz CW signal. And no errors. Yay!  ;D
1012666-0
2. Local FM station
1012662-1
3. 2.75 GHz to 20 GHz sweep test
1012658-2
« Last Edit: June 29, 2020, 12:33:15 pm by 1Ghz »
Offline shakalnokturn
Did you open the YTO assembly to see if anything could be done to it or just for the sake of autopsy?
  • Country: kr
Bad YIG oscillator photos.
Inner circuit
1013614-0
1013602-1
1013606-2
1013610-3
1013582-4
Coils
1013618-5
1013586-6
1013590-7
Maybe FM coil is burnt?
1013594-8
1013598-9
The following users thanked this post: shakalnokturn
FM coil does look a bit suspicious.
I'm not sure what's been going on here either, looks as if it has heated enough to change the gold plating's colour.
The following users thanked this post: 1Ghz
Online dietert1
  • Country: br
    • CADT Homepage
Two years ago i fixed a similar problem on our HP 8560A. As far as i remember the YIG oscillator uses a special 28V voltage that comes from a voltage doubler. Those voltage doublers were made with electrolytical caps and those don't last as well, since on power-on they get reverse voltage every now and then. Replaced one cap and the spectrum analyzer worked again.
The instrument is servicable, yet the power supply is buried deep inside, so it takes patience to get the board out. The capacitor has 3 pins with the third pin connected to a built-in shield foil. Made something similar from a normal cap and copper foil i had.
Regards, Dieter
  • Country: kr
Yes, it's suspicious! Maybe too much current? Hmm...  :-//
If it fails again, I'll check the caps.
Offline romal76
  • Country: ru
Re: [SOLVED] HP 8563A Spectrum Analyzer Repair - Is the YIG oscillator dead?
« Reply #11 on: January 02, 2022, 11:34:05 am »
I'm having 8561E with bad YIG (p/n 5086-7906).
Is it possible to replace bad YIG 5086-7906 with another one 5086-7903? Which steps? How to make tuning after replace if it possible?
  • Country: kr
« Reply #12 on: January 28, 2022, 08:01:04 am »
Sorry, I have no idea. I replaced it with same YTO.
Offline romal76
  • Country: ru
« Reply #13 on: January 28, 2022, 09:07:57 am »
It's working good, I was check it. 5086-7906=5086-7903
Next problem for me is SYTF, broken switch.
  • Country: kr
« Reply #14 on: January 28, 2022, 09:41:16 am »
Nice to hear that. Thanks for share your experience.  :-+
Hope you can fix it soon.  :'(
Offline larry540
I am new to this forum and recently purchased a 8561e which 28v supply is running at around 7v. It turns out if I unplug the SYTF(5086-7803) the supply goes back to 28v. The SYTF looks to be pulling well over 800ma on the Tune input. I don't belief this is normal. Does anybody have any comments? Is there any hope of repair?