EDIT: June 29, 2020
TL;DR
The problem was caused by bad YIG oscillator.
Replaced the YTO (YIG oscillator) and that fixed the problem.
I recently bought an used HP 8563A spectrum analyzer.
But it's not working and it displays an error messages.
301 YTO UNLK
317 FREQ ACC
334 LO AMPL
Here is the pictures.
1. 8563A 1st LO Output (picture from the 8563E)
[attach=1]
It's not working correctly. No signal output from low to mid frequency range.
2. 8563A CAL Output (picture from the 8563E)
[attach=3]
It's working correctly.
3. 8563A Power Supply LEDs
[attach=2]
I think it's working correctly.
Is the YIG oscillator dead? Should I replace the YIG oscillator?
Thanks,