Author Topic: HP 8566B spectrum analyzer trace flicker/glitching  (Read 940 times)
Offline tonycox01Topic starter
« on: January 14, 2020, 10:06:54 am »
I recently got hold of an HP 8566B SA which largely seems to be working OK apart from a random flicker/glitching of the trace display. The trace appears to break down in to a flurry of horizontal lines at random intervals and positions. I've trawled for similar issues with the 8556B but not come across anything so I thought I would ask a question here before diving in to serious troubleshooting - just in case anybody has seen a similar problem or has any suggestions about where to focus my troubleshooting efforts.
I have created a short video that demonstrates the issue (this brief capture was from the self calibration routine).
https://youtu.be/LvoDBJFQfnM
thanks in advance for you help/suggestions.
Tony
Offline jfphp
Re: HP 8566B spectrum analyzer trace flicker/glitching
« Reply #1 on: January 14, 2020, 10:11:20 am »
I will try with the power supplies (ripple) mainly those going to the display board.
Offline tagchen
« Reply #2 on: January 14, 2020, 10:28:37 am »
in this old measurement tools check power supply caps, mostly higher voltage (400-450V) are the problem.
this problem can be keramic caps for digital chips (cpu boards or so)
have you picture from inside?
installing retrofit document can little help.
* 08566-90151-retrofit.pdf (4410.18 kB - downloaded 101 times.)
service and repair manual to big to post, find in keysight page
Part Number: 08566-90042 (Dec84). This manual is provided for information only. The 8566B and 8568B are no longer sold by Keysight. Note: the file size is 36.37 MB.
« Last Edit: January 14, 2020, 10:34:41 am by tagchen »
« Reply #3 on: January 14, 2020, 01:40:35 pm »
Thanks - no picture yet but I'll try and get some.
I'll start with the caps on the power supplies and see where that goes.
cheers
Tony
Offline Johnny10
« Reply #4 on: January 14, 2020, 02:39:10 pm »
I might check for input power line noise.
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B