Electronics => Repair => Topic started by: Daxxin on October 15, 2017, 07:52:53 am
While CAL FREQ is done without problems, with CAL AMPLTD attempting i get ''CAL: RES BW AMPL FAIL'' service manual
says require realignment of crystal and LC bandwidth filter at chapter 2 , boards A11 and A13
but following the instruction look like there is errors . Wich is the correct sequence about jumper?
(http://oi68.tinypic.com/2ajxffr.jpg)
For which sa are these instructions? They differ between 8590a, 8590b and subsequent versions.
it's about 8590E series , there no errors i didn't get that need to put all three jumpers at start of procedures
not one per time |O that's my fault , anyway i'm still having trouble with CAL OUT seems its not enabled
when i switch on sa , and even if i try to CAL AMPLTD i get 'CAL SIGNAL NOT FOUND'
(http://oi67.tinypic.com/2iiy7eo.jpg)
If i start CAL FREQ fisrt then the CAL OUT signal appear ..why?
(http://oi68.tinypic.com/ot3tpd.jpg)
anyway it look not good enough to complete the CAL AMPLTD.
According with the manual settings for Crystal and LC alignment (freq 300Mhz , span 200Khz , BW 30Khz , AMPLIT. -20dBm)
when jumpers inserted signal appear worst.
(http://oi68.tinypic.com/50flzc.jpg)
Following this steps found in another forum everything ok :-+ CAL AMPLTD and FREQ completed and stored
Preset
Frequency -2001 Hz
Amplitude, More 1 of 2
Input Z 75-ohm (50-ohm in my case)
Amplitude, Scale Log Lin (Log), More 1 of 2 Amptd Units dBmV
Amplitude, Scale Log Lin (Lin), More 1 of 2 Amptd Units dBmV
Cal, More 1 of 4, More 2 of 4
Service Cal, Stor Pwr On Units
I would then default the calibration data by doing the following:
Preset
Frequency -37Hz
Default Cal Data
then CAL routine.
I am having the same problem with 8596E. How do you make those crystal short jumpers?