So my HP8590A works as a spectrum analyzer but I caught it being funny a few times before.
1) i was looking at signals and for some reason a bunch of TV-band frequencies turned into a noise for some reason.. as if the SNR of the instrument suddenly got horrible until it was reset, so it looked like a clean but noisy display wheras before there was quite a few peaks there
2) The mixer level adjustment seems to 'latch'. If I set the dB/division low (say 1-2 dB/div) and I scroll the mixer too hard (i.e. the signal trace goes off scale on top of screen), then the mixer level cannot be moved back up, and I need to hit the reset button.
I thought that #2 might be something to do with a bad DAC? I am not sure why it would work fine moving the mixer level up and down fine, but if the signal goes off screen the mixer knob becomes unresponsive.
I got this instrument at a bargin price so there is probobly a fault with it.... but I am not even sure if its a memory problem, some kind of intrinsic problem with the analyzer, or some kind of analog problem with 'latch up' of some kind in the DAC..
So I turn on the instrument, turn on attenuation to 0dB in amplitude screen, put on 2 db / division, using keypad, so the signal is under the bottom of the display. Then I press the reference level button and adjust the trace using the knob, until the trace is shown on screen. Now, when I adjust the knob,m the number on the display changes but the trace does not move. If my velocity on the knob is high, then I can totally hide the trace before it 'locks control out'. If I click the reset button, it works fine.
But if I don't play with the attenuator setting, when I change the mixer level, I hear relays clicking and it never locks out, but it changes the attenuator level. It only locks out if I first set the attenuator. But if the trace gets frozen, I can move it up further, just cannot move it down.
Is this a intrinsic property of this spectrum analyzer? I thought I caught another one doing something like this before but I don't have that much experience with it. It feels like there are forbidden zones of operation. My more analog spectrum analyzer has some kind of rotary linkage that makes sure the knobs have the correct states in reference to each other, so you can't run it on silly settings.
But I do find it weird that the system locks out instead of warning you or whatever... (it feels broke to the user if you can suddenly only further hide a trace/have a lying display, since clearly the mixer level is not changing anything despite the numbers changing) Can anyone confirm with a 85xx instrument with a hardset attenuator? You don't need a signal or antenna for this, a bare instrument will do. If this is a design, they really should have put a 'invalid setting' notification pop up on the display.