I just purchased a problem unit HP8590A Spectrum Analyzer it has measurement problems below 17MHz.
I have read the service manual and done the software tests but before opening the unit I wondered if there was any advice from those of you who have worked on unit.
No errors listed or in memory
Manual says problems are most likely in the FM coil circuitry.
So what I have done so far.
Done a Cal Frequency and Amplitude cal and store.
Run the FM Coil Drive Routine
Pictured
Done the ADC routines.
Seem normal switches at listed frequencies.
Checked LO for Frequency
0 2.050901 GHz
16.92MHz 2.07947 GHz
1.5 GHz. 3.548310 GHz
Measurements and Amplitudes
Seems I get wrong Frequency measurements from 17MHz down.
I get wrong Amplitude from 2MHz down.
Drifting signal (Where peak walks across screen) from 200K down.
Amplitude drops pretty consistent -60 dBm
I also noticed as I worked the Amplitude on the signal generator the Amplitude measurements in 8 to 2MHz intermittently were correct.