Hi everybody
I recently got a HP 8594E spectrum ( not logic

) analyzer off ebay from south Korea at a reasonable price. Needless to say it was sold as defective with no further infomation.
The front panel isn't in good nick , the brightness knob is broke , many scratches and faded inscriptions and a missing screen filter.
But it powers up , the YIG osc is fine and displays the calibration signal at -30 dB instead of -20dB. The freq calibration passes but the amplitude calibration fails with a RES BW AMP FAIL error message.
After some RTFM I found out how to cablibrate the A11 and A13 BW filters. So I made three XTAL shorts with 10nF and 90 ohms in series and carried out the procedure. It's very easy to do and take 30 minutes to do it properly.
After that the AMPD calibration passes every time

Buuuut not the confidence test...
And here is my problem : the conf test fails at the very begining with a POS PK FAIL message. The analyzer seems to be working fine tho.
According to the SM the processor board or analogue interface is defective.
Did some measurements with my scope at the input of the DAC on the processor board and the signal looks fine. the 3rd converter DAC's correction value is 199. A bit high but no problem whatsoever.
Then i noticed that if I set the ref level to -58dBm bad things happen ( see photo ) same thing at -68dBm and -78dBm. everything is fine below -57dBm
The default doen't appear on my scope tho.
I don't have a swap processor board but I swapped the two A11 and A13 BW filters 3rd converter board and the A12 attenuator board and the failure is still there even with the 321.4 Mhz input disconnected.
so I suspect the positive peak detector on the processor assy but I can't figure out how it actually works (got the CLIP also )
Does someone have an idea abt how to solve this ? Or give me some informations on the peak detector ? What does it do exactly?
Regards