Author Topic: HP 8596E repair, Freq Cal pass, Amp cal fail on low signal  (Read 998 times)
Offline lfsrougerTopic starter
Check the cal signal, looks normal w/o correction, distort w/correction. Could be a wrong calibration in past? Any advise?
Online MarkL
Re: HP 8596E repair, Freq Cal pass, Amp cal fail on low signal
« Reply #1 on: May 06, 2021, 04:47:06 pm »
The first thing to try is defaulting the cal data and re-running the self-cal.
You will need to enter a passcode to load the defaults.  The procedure is described in the User Manual and also in the Assembly-Level Repair Manual.  Search for "default cal data".
And while you're at it, if you haven't backed up your correction constants yet (this is not the same as the cal data), I recommend performing the backup procedure in the Assembly-Level Repair Manual (chpt. 3).
« Reply #2 on: May 06, 2021, 07:22:17 pm »
Thank you! Will try as suggested.