Here is a fascinating story about a tiny capacitor that took down a monster $$$K device. I did this repair some time ago on my 8753C VNA.
VNA all of a sudden stopped working. Could not power it up anymore. It looked as if the protection circuit kept triggering on one of the secondary voltage supply rails. After troubleshooting the problem was traced to the power supply which stopped producing the proper +5.1V voltage.
Inside the power supply. The left half is the power stage and rectifiers. The right half is line input and the controller circuit, which was of my interest.
I had to reverse engineer the PS unit a bit around the controller IC - see a fragment of the schematic below . To my surprise, found a 0.22uF bypass capacitor developed a leak, corrupting the reference voltage on pin 2 (the controller IC error amplifier). That caused the power supply +5V channel voltage change erratically between 4.5 to 5.1V which apparently triggered the under voltage protection circuit on the 5.1V supply rail which in turn was shutting down the whole power supply.
And here it was the bad guy... darn penny cap that was causing multi $$$K pain in the arse
The bad capacitor was replaced with a mica 0.22uF new one. After the replacement the +5.1V voltage became stable within 0.1mV and the problem was gone. The VNA powers up normally now.