I did some probing around tonight.
I started by going thru the HP trouble shooting procedure for the oscillator board.
First I noticed that the output at the first oscillator (first attachment) is 4.70 volts, it is supposed to be 3.00 Volts RMS. The P-P voltage is 12.64V. Looking at the schematic (second Attachment) the zener diodes on the output of the U4 could be clipping the output causing the harmonics. If that were the case it would seem the voltage would be 10.2 Volts instead of the 12.64V.
So I'm thinking something is wrong in the ALC (Automatic Leveling Circuit) somewhere to make the output so high.
I found the first problem at TP3 which is the output of the first sample and hold circuit. The 3rd attachment shows the waveform at TP3. The magenta trace is TP3, the yellow trace is the gate of transistor Q15. The 4th attachment is from the service manual and shows what TP3 SHOULD like like.
See the schematic of the first sample and hold circuit in the 5th attachment.
This a fairly simple circuit. There's not much that can go wrong. The Op amp, 2 caps, and Q15. I'm not very good and circuit analysis like this but I'm thinking Q15 is either shorted? or open? The op amp is working, I don't see how either of the caps could cause this behavior.
If anyone has any thoughts I would love to hear from you.
Gary