| Agosto:
I posted before about ongoing issues I am having with my newly acquired 8903A: After sorting out relay issues on the Input Board I tested the basic functions of the unit. Selecting 1000Hz, 3V AC I connected High Output with High Input and measured distortion. "60-72dB" at 1000Hz, getting progressively worse down to 20Hz with 35dB (readings not stable). Looking at the waveform at the High Output with a scope I noticed a distortion at the zero point of the sine wave fading away at around 500Hz. Looking around the Oscillator Board I re-soldered a portion of the board to no avail. Using an external signal source (1000Hz, 3V AC) I am getting 76dB on distortion measurement. So I am quite sure the issue is somewhere around the Oscillator Board (does 76dB sound ok in terms of performance for a $500 signal generator, GwInstek??). Meanwhile, I got my hands on another Oscillator Board and used that instead: Upon initial start-up the distortion level improved to spec (around 87dB) but after around 2min and some faint (REED?) relay clicks distortion suddenly changed to the original board's level (60-72dB). So I am afraid I damaged the new board. Supply voltages are all within spec. What puzzles me now is that the unit worked fine with the new Oscillator Board, although only for a short time. Am I wrong in thinking that the problem may originate somewhere else? I am not experienced in servicing HP equipment so I hope someone here on the Forum can point me in the right direction. Thanks!! |
| wn1fju:
After the two minute warm-up and things go astray, does the distortion measurement change when using an external generator? In other words, can you isolate the problem to the generation side or the measurement side. -76 dBc is a bit optimistic, in my opinion, for cheap signal generators. Most of mine are in the low -60's, but occasionally you will find an oscillator that uses a Wien bridge circuit and those can easily top -80 dBc distortion. If you have a function generator, you might see what the 8903 measures on a 1 kHz square wave and triangle wave. You should get approximately -7 and -18 dBc respectively. Also, when you say the power supply voltages are in spec, was this after the thing started to misbehave? Have you looked at the power rails with a scope to make sure there is not excessive ripple? Sometimes, measuring DC voltages with a voltmeter is not enough since it won't respond to AC ripple and just gives you an average amount. I would be surprised if you damaged the other oscillator board. HP's designs were usually pretty robust. There is nothing "special" about servicing HP equipment of the 1980's era. Unless, of course, you understand that HP had superior engineers, produced superior equipment and published superior service manuals. |
| Agosto:
Thank you for your feedback - much appreciated! To your questions: 1. Yes, using an external signal generator (the GwInstek I was referring to in my post) distortion measurement improves from 60-72dB/1kHz to -76dB. at 20Hz, improves from approx. -35dB to -76dB. So I concluded the root cause is the internal oscillator. 2. I checked the P/S voltages with a scope: No excessive ripple. The +15V rail is a bit low (14.7V) but as far I can read from the service manual that is still within spec. For clarification: The new board showed excellent distortion numbers after my first use of the board. After the values went out of spec, measurements are consistent. There is no more spec ratings prior to a "warm-up phase". The bad readings are now consistent, no matter if cold or warmed up. The change from in-spec values to the lower values happened instantly, so no gradual degradation. However: (!) Today, after 2 days off, I started the unit to measure square (Result: -7dB) and triangle (Result: -18dB) distortion rates. To my surprise, sine measurement at 1kHz/3V showed -83dB distortion again! So something is not stable... I did a number of measurements at different frequencies, with internal oscillator and external signal generators at 3V. Here are the results: Frequency Int. Oscillator Ext. Generator 1kHz 83.5 75 500Hz 83 75 250 79 75 100 71.5 75 20 58 75 10kHz 83 75 50kHz 88 78 100kHz 89 78 Although for unknown reasons the measurements are now better than what I originally reported, I believe that the results below 1kHz e.g. -58dB at 20Hz are still outside of spec. Correct? Should I check supply voltages again? |
| wn1fju:
For reference, I set my 8903A with the high output connected to the high input, no termination, 3V amplitude and dialed in your frequencies. My unit defaults on power-up with the 80 kHz LP filter enabled, so that's how I left it. Both front panel float switches were set to the ground position. I got between -90 and -95 dBc at 20, 100, 250, 500, 1K, 10K and 50K frequencies. With the 100 kHz frequency, I turned off the 80 kHz LPF filter and got -76.6 dBc. With the LP filter off, the readings at the lower frequencies were all about 2 - 4 dB worse. And while I'm thinking of it, you might want to reseat all the boards and give all the pots a jiggle (after you have marked their positions). Perhaps a pot is dirty and needs a little jostling. Also, give all the boards a good rap and see if anything on the display changes a lot. The notch filter tune and balance adjustment is particularly sensitive in my opinion. They use a single-turn pot and you had better have a very steady hand! In fact, on my unit, I actually replaced that pot with a ten-turn variety so I could really dial in the lowest distortion. |
| Agosto:
Thanks! Meanwhile, I was trying to do exactly what you described by wiggling the boards around and reseated all boards. Nothing - no value changed. But: The moment I switched the High Output float switch to float the 1kHz value came down again to 70dB. Any attempt to bring it back to 84dB by switching again, wiggling the switch or even shorting the switch on the back was not successful. I will try the pots as you suggested. Other than that I am at a lost (at the moment). Thanks again for your support! |