I've got an HP 8903A that has been working fine for a year or 2. I've got several threads here where I have repaired other problems. This is one of 2 units I have so i'm not sure which problems this one has had in the past but It' been working well.
One day I was checking an amplifier and the AC voltmeter seemed high. I checked and it was.
When first turned on it's about 16% high and then as it warms up it goes to about 25% high. It seems to be across all the metered values and all ranges. I have focused on the AC voltmeter.
I have been working on this instrument for about a year on and off and I don't know where to go next. I need help.
First I'll go thru what I have done.
The oscillator is perfect, the frequency is accurate and the output voltage is accurate.
As mentioned above I have 2 of these and the other one works fine. I swapped cards in and out until I found that the problem is on the A4 Output Amplifier/Voltmeter card. So the problem follows this card from unit to unit.
First thing I did was to check all the lytic caps and all the white axial Caps (Film?). They all tested good on 2 different capacitor checkers.
I went thru the troubleshooting procedure for the voltmeter (Service Sheet 7) and everything checked out. I couldn't find anything that looked suspect. All the waveform examples looked good.
See the attached schematic for the voltmeter section.


Since the meter is high on all functions I reasoned that the problem must be after the U18 Mux. It checks out, the output is always equal to whichever input is selected.
I checked all the voltages that are called out on the drawing. They are all good.
I don't understand totally how this circuit works, but the description given in the manual is;
All values are changed to a DC voltage and presented to the + input of a comparator U20
The computer generates a ramp gate signal, that starts a counter counting in the computer.
The gate also starts a ramp voltage (starting @ about -6V) at the - input to the comparator.
When the ramp voltage equals the DC voltage on the + input the comparator turns on.
The comparator output +5V is used to signal the computer to stop the counter.
The higher the measured voltage the longer the ramp runs and thus the higher the count and the measured value.
See the attached O'scope shot of the comparator output (top trace in yellow) and the ramp signal (bottom trace in magenta)

They look good to me, but a 25% difference might be hard to detect.
I thought that the problem would most likely be in the ramp circuit, around U15.
I started checking all the components around it.
Since C70 controls/generates the ramp I thought it might be bad or going bad. I had already checked it and it was good. I checked again and it was still good. I replaced it with a new one and nothing changed.
I checked Q11 and Q12 and they are good.
I checked Q9 and it is good.
I checked VR5, the voltage on it was 5.85V instead of 6.2V. I installed a new one with a voltage of 6.2V, but no change in the meter reading.
This is where I'm stuck, I feel like I'm missing something simple.
I don't know what to check next.
Any ideas from a fresh pair of eyes would be great!
Thanks
Gary