Quoted here from another thread, to keep the discussion here:
I dropped my 2378 DVM and a part labeled PTC became disconnected. The markings are TDK 9309. I purchased a non working 2378 and removed the TDK park. This one had markings TDK 9009. I installed the 9009 into my working 2378 but it is not installed correctly. One of the lead ends became unsoldered and the repair is not very well done.
The fault I had with the missing part was in the Resistance function. It would only read OL or 0.00. Now it reads correct resistance values, I tried a 500 ohm resistor.
I would like to install correctly a new PTC but I am unable to find a replacement.
Anyone have a part number and a supplier for this part. I think it is a temperature controller resistance.
Regards,
Dan Kahn
You need to know the PTC cold resistance. You could measure the resistance end-end of the piece that broke.
Today 1,100Ω is commonly used by Fluke and 1,500Ω ohms in the low cost chinese meters. To give a ballpark value. It's not critical. Out in the wild I see 500-3,500Ω used.
Pics show two styles of PTC in your meter, a long square part with no epoxy coatings or a black disc in the Mr. Modemhead pics.
From the user manual, max. voltage. rating on Ohms is 500V so a 600V PTC looks reasonable. There are few PTC's that are rated for 600-1,000V on the market, only two manufacturers of decent parts that I am aware of.