Hi,
Does anyone have any experience with the HP /Agilent / Keysight E3630A Power Supply and can confirm if this model "normally" displays 0.01A on the Amps display for the 6V meter setting (no load, case cover on, some warm up time), instead of 0.00A as one would expect?
I picked up a reasonably priced unit on ebay that seemed to just needed calibration. So I followed the meter calibration steps in the service manual, using my calibrated Fluke 87V. Everything with the supply checks out OK (e.g., no oscillations, outputs voltages agree with the display), except the 6V output Amps meter displays 0.01 amps with no load. The +20V and -20V meter settings display 0.00 amps with no load as expected.
The 6V output does display the correct current with various loads up to current limit, so the meter calibration appears to be OK. But...I found the 6V and 20V Amps meter calibration steps interact with each other, so I can't zero the 6V Amps display without affecting the calibration of the 20V Amps display.
Thanks,
Richard.