I've got an older HP branded 54845A which was failing several tests and all calibrations. I have fixed everything up. Replaced the hard drive with a compact flash, fresh install of the software and all the upgrades up to V4.5 and added memory (64MB now from 16MB before). Cleaned up the ADC ASICs and replaced an on board regulator chip and a couple of capacitors (on acquisition board) and now the scope passes all self tests (multiple times), passed all calibrations and everything is honky dory except a peculiar issue that was also there in the beginning when I got the scope (it was win95 and SW V3.72)
As you can see in the pictures, when the graphic display is enabled, all is fine and the scope works perfectly as far as I can tell. As soon as I disable the graphic display (top right corner) the waveform is cut into pieces as you can see. The issue is very visible on fast time bases but it also exists on slow time bases specially with low frequency square wave it is very visible.
For the life of me I cannot figure out what can be wrong with it. I don't think it is the acquisition board. When I got the scope I thought it was a software/driver issue so I replaced the clunky noisy hard drive and upgraded the software but the problem persists. The only hardware issue I can think of is the display board on the PCI slot but what is in there to test?
Can someone with this model scope please check and see if the scope shows the same waveform when the graphic display is disabled?