I just a 1630D from a user on the forum and it worked great except for inability to sample under 500ns. Inside the unit there are several boards in a card cage, with all of the logic and cpu cards underneath the power supply card, which makes them inaccessible. The bus plane the cards plug into is mounted vertically at the front of the device. There is a bus connector at the top mounted parallel to the top cover of the unit. So, if you need to trace a problem on a board, you can pull it out of the card cage and plug it into the service port on the top side of the machine. Neat!
The problem with this machine was a stuck 200MHz oscillator circuit on the master timing board. I followed the procedure in the service manual and got it working again. (Just had to adjust a variable capacitor). Two of the flip chip ICs on the timing master board get ridiculously hot and have to be kept cool. I put a fan on a stack of junk on a char to keep air rolling over the heat sinks. I also attached a sensor on one of them to monitor the temperature while I carried out the repair.
Here's a link to the service manual:
http://bitsavers.informatik.uni-stuttgart.de/pdf/hp/te/01630-90917_1630_Service_Apr88.pdfNot a particularly useful post, nor an interesting repair but I think the attention to detail HP put into the design of this unit is nice.