Hello,
last year I acquired both a HP34401A and a Instek 8251A and not used them much.
This year I got a DMMCheckPlus that would give calibrated VDC, VAC, DCV and ACI.
The values are 5VDC(± 0.007%), 5VAC(±0.1%), 1mA DC(±0.1% ), 1mA AC(±0.2%) plus 100k, 10K, 1K, and 100Ohm precision resistors(0.1% - 10ppm). It came with a paper with actual values checked against a traceable source.
Not a full professional calibrator, just to have a quick check about a DMM health.
Most of the stuff seems "okay" on both DMM:
5VDC
Instek Hp
4.9991 vs 4.99996
5VAC
Instek Hp
4.9935 4.99346
1mA DC
Instek Hp
1.0000 vs 1.00003
except when I test 1mA ACI:
On the 8251A I read 1.0964mA (and 0.37mA with probes unconnected, may need calibration as well)
On the HP34401A instead, I read 0.12mA when unconnected, and basically reading does not change if I connect the 1mA ACI source.
If I use the 1mA DC source, it correctly shows 1.0001mA
I tried to understand IF this DMM could measure below 10mA AC and apparently, it can't (error on 1A range is 0.1%) so nothing wrong here.
According to this, this DMM does have a hidden 10mA range(but not a 100mA one!), so this is another incentive to buy a HPIB card:
https://www.keysight.com/main/editorial.jspx?cc=IT&lc=ita&ckey=1000001267:epsg:faq&nid=-31895.536880933.02&id=1000001267:epsg:faqCould anybody please confirm it? Which is the lowest AC current I could expect to measure reliably with the HP34401a?
I would like to sell the 8251A, thinking one HP34401A would suffice and was way better...now this weird feeling that the Instek does something the HP doesn't....
TIA
Giuseppe Marullo
IW2JWW - JN45RQ