I have an HP53132A that works very well except for one thing. It won't hold calibration.
This is an older machine with firmware 3403. OCXO is an older but original HP board. What's weird about this board is there is no HP11801 socket but there are holes on board that will take different OCXO other than Isotemp that medium standard. By tracing the line, I've determined it is compatible with MV89A. I stuck one in, performed a cal, it cal'd correctly.
In the man time, I have a splitter on house standard and an independent GPSDO that acted as a standard for calibration purpose. They were fed to an oscilloscope. When calibration ended, they were correctly in sync. Traces are stable. Nothing wrong. Now, I press FREQUENCY 1. Suddenly, it loses calibration and I can see traces on scope losing sync.
I suspected EEPROM has a dead spot at the exact location where OCXO EFC data are kept. But calibration count is only 30. While it is not impossible, it would be hard to imagine someone went through trouble to reset this.
Are there any LOGICAL reason how else this may happen?