Author Topic: Fluke 8050A decimal point problem  (Read 1481 times)

0 Members and 1 Guest are viewing this topic.

Offline mellowamoyaTopic starter

  • Newbie
  • Posts: 5
  • Country: tt
Fluke 8050A decimal point problem
« on: April 20, 2017, 03:51:25 pm »
Hi All,
Im hoping someone can help with this my trusty Fluke8050A that has been on the bench for many years and worked very hard, it has developed a problem, it still works and reads but all the decimal points on the display are on all the time no matter what button is pressed, sometimes the last one on the right will be flashing but most of the time they are just all on.
any pointers as to fault find the problem would be great.
thanks
dennis
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 8050A decimal point problem
« Reply #1 on: April 20, 2017, 05:26:10 pm »
8050A Service manual at

http://bama.edebris.com/download/fluke/8050a/8050a.pdf

suggest for (from table 4-13 troubleshooting guide)

All or no decimal points that U16, U17, interconnect may be at fault.
« Last Edit: April 20, 2017, 05:28:13 pm by retiredcaps »
 

Offline JFJ

  • Regular Contributor
  • *
  • Posts: 236
  • Country: gb
Re: Fluke 8050A decimal point problem
« Reply #2 on: April 20, 2017, 05:27:11 pm »
The decimal points are driven by U16 - a four segment LCD driver (CD4054BE). Hopefully, replacing that IC will cure the fault. If not, then it may be game-over, because the decimal point control signal comes directly from pin 22 of U17 (Fluke's own 40 pin controller).

If your meter is the 8050A-01 (rechargeable battery option) then the display will have a 'BT' (low battery) indicator, which is also controlled from pin 22 of U17. So, if 'BT' isn't permanently on, then U17 is unlikely to be faulty.

Sorry, it's the 'REL' indicator that is "also controlled from pin 22 of U17" (I misread 'Out 4' as pin 4 on the CD4054B datasheet). So, if 'REL' isn't permanently on, then U17 is unlikely to be faulty.
« Last Edit: April 20, 2017, 06:41:04 pm by JFJ »
 

Offline ModemHead

  • Frequent Contributor
  • **
  • Posts: 731
  • Country: us
  • No user-serviceable parts inside.
    • Mr. ModemHead
Re: Fluke 8050A decimal point problem
« Reply #3 on: April 20, 2017, 11:46:30 pm »
Before heating up the soldering iron, disassemble the ribbon cable interconnect and give it a clean, just in case.
 