apart from my tube problem…..there is a post i found (see below)
i have the same series errors like 355, 351, 353, 319, 317, 356, 335 and the same sympthons....
because i see not input on the screen i did a quick check on the direct output of the YIG oscilator it give me arround 5.780Ghz at about +10db.
this post is about a 8560E, but many boards are the same
The fault is the analyzer will not display any input signal including the calibration input signal. You will also have numerous errors like 355, 351, 353, 319, 317, 356 and 335. Your mileage will vary. Might have less or more errors.
In 8 months, I bought two different 8560E analyzers off of ebay from different sellers. Once I received them and checked them out, both had the symptom listed above. I bought the second one because the display was still nice and bright compared to the first one I purchased and it was cheap. These analyzer are known for having dim displays. Anyway, both units had the same symptom of no input signal showing on the screen and it turned out that both had the same faulty part.
Bottom Line: Go to the A15 board and look at the J701 output. Should be 600Mhz AND around -3dbm to +5dbm. Both of my units were spot on 600Mhz BUT the levels were around -20/-25 dbm. After some troubleshooting, the problem was U700.
After replacing U700 on both units, the errors went away and the analyzer worked as it should. U700 is just a RF amp on the A15 board. It's under one of the metal shields/heatsinks in the 100Mhz distribution section. The 100Mhz gets doubled, then tripled to 600Mhz.
U700 is a MSA-0505. I purchased genuine HP parts from a seller on ebay from Israel. There are others from China but who knows if they are genuine.
Is this fix also related to the 8563E, to call it by its name: did someone ever use this fix in a 8563E?
all tips to check and isolate my problem are welcome..
help on how to easy check my
- 1st mixer (5087-7032)
- low dis. ampl. (5086-7885)
- 2e conv (5086-7957)