Author Topic: HP8590L spectrum analyzer calibration signal issues  (Read 4826 times)
Offline BjoernTopic starter
  • Country: sg
« on: July 10, 2018, 06:16:05 am »
After I replaced two capacitors on the analog board I got the unit up and running. However the calibration routine seems to have issues with finding or using the calibration signal.
I made a YouTube video showing the issue and would very much appreciate your input:
Many Thanks !!
Offline borghese
  • Country: si
Re: HP8590L spectrum analyzer calibration signal issues
« Reply #1 on: July 10, 2018, 06:30:08 am »
Can you specify which capacitors have you changed?
Maybe some loose connector? The calibrator output is -30dBm.
Cheers
Borghese
  • Country: sg
« Reply #2 on: July 10, 2018, 10:44:19 am »
Hi and many thanks for the prompt reply !
I do not have a schematic of the instrument therefore its a bit hard to say which of the caps I changed without opening it up again.
When I bought it the +15v were down and I found that on the analog board two of the tantalum caps were shorted. I assume they are filtering the main 15v supply of the board. When I changed them to electrolytic the SA powered up.
Do you know on which board the calibration signal is generated ?
Would I have to find the calibration signal that is accessible from the back or the one from the front panel ?
  • Country: sg
« Reply #3 on: July 10, 2018, 11:46:08 am »
Luckily I found the schematics within a large manual.
I remember the Caps to be C40 and C50 (see schematics) which seems to be on part of the board that might cause the error.
Maybe the OP-Amps did suffer ?
What do you think ?
Offline SaabFAN
« Reply #4 on: July 11, 2018, 12:18:05 pm »
I think your general idea where the problem is (Yig Coil-Drivers) is correct. An unmodulated signal shouldn't look like it does on your screen. Which means either the coil-driver itself, or the control circuits for the driver are bad / horribly out of calibration. Have you tried adjusting any potentiometers or measured them? What do the control-voltages for the YIG look like?
Btw. How are the Supply-Voltages doing? Voltage ok? Excessive Ripple? If one Tantalum has gone bad, there's a chance others are causing problems as well.
Also with the Cal-Signal supposed to be -30dBm, it's quite a bit lower on your screen. With strong harmonics (don't know if they are supposed to be there). I'd suggest you build a simple diode-detector (if you don't have any equipment to measure the cal-signal) and measure the output-power. Doesn't have to be precise, just a rough estimate to see if the signal is in the right area.
« Last Edit: July 11, 2018, 12:23:01 pm by SaabFAN »
The following users thanked this post: Bjoern
  • Country: sg
« Reply #5 on: July 11, 2018, 12:36:00 pm »
I am just a few days away from finishing my Signal Tracer project.
Its based on Carlson's Super Probe (Youtube) and I hope that it will help me find some faulty cap's.
Offline borghese
  • Country: si
« Reply #6 on: July 11, 2018, 08:19:53 pm »
The level of the calibration signal is -20dBm; my apologies.
Your analyzer settings do not allow you to view the 300MHz calibration signal; the start frequency is 400MHz; there is also a spurious signal at about 410 MHz.
Can you post a photo at full span?
This is my analyzer's screenshoot
« Last Edit: July 11, 2018, 08:21:40 pm by borghese »
Cheers
Borghese
  • Country: sg
« Reply #7 on: July 12, 2018, 02:16:32 am »
Hi attached is my screenshot, does it give any clues of were to search ?
I hope that my BNC cable connection on the back panel (RF in/out) is acceptable ?
Offline borghese
  • Country: si
« Reply #8 on: July 12, 2018, 05:32:52 am »
Hi
Have you a signal generator for testing the frequecy response? Also the local oscillator feedtrough is very low, check the amplitude and stability of LO. First check all power suppy.
« Last Edit: July 12, 2018, 05:34:59 am by borghese »
Cheers
Borghese
  • Country: sg
« Reply #9 on: July 14, 2018, 09:37:25 am »
I checked all the voltages which are OK.
I checked the cal output on the backside which gives a nice 10Mhz signal and is OK as well.
Attached are more screenshots of the calibration process, I hope that it gives further clues what might be at fault.
  • Country: sg
« Reply #10 on: July 14, 2018, 11:44:54 am »
I am thinking, that maybe not the calibration signal has an issue but the input of the signal analyzer.
I connected my Signal Generator and fed a simple 10 Mhz none modulated signal into the SA input.
Attached are the screenshots which show that the unit definitely has issues.
Probably these issues do not allow a proper calibration, or do you think that the misreadings are due to a failed calibration ?
Where could I start searching, are there modules I could skip (inside) in order to sort of trace my 10 Mhz signal ?
Offline borghese
  • Country: si
« Reply #11 on: July 14, 2018, 02:38:54 pm »
Can you post a screenshot with start -5MHz e stop 45MHz and center 10MHz, span 100khz and RB 3khz? Also you have a lock off message
Cheers
Borghese
  • Country: sg
« Reply #12 on: July 14, 2018, 02:57:22 pm »
I tried to match your request as good as possible but some settings change others.
I start by setting the center f to 10Mhz when I set the start -5 and end 45 then the center has changed.
The attached is a close match.
What is the RB 100khz setting I cant find it (what does RB stay for ?).
I noticed the Lock Message but I am not sure why it is there or how to get rid of it.
Hope the photo helps identify the error.
Highly appreciate your support !!!
  • Country: sg
« Reply #13 on: July 14, 2018, 03:27:37 pm »
Sorry I noticed that the reference BNC cable on the back was disconnected.
I did the same test again with the signal connected at the back.
Offline borghese
  • Country: si
« Reply #14 on: July 14, 2018, 03:46:21 pm »
The signal dont look bad.
Start -5 stop 45 is the same as center 20 span 50. RB is resolution bandwith under the button SWP/BW.
New setting:
Centre 20 span 50. AT 50 dB; take a shot.
Centre 10M, span 100khz; screenshot. Wait 60 minutes or more for a warmup
You can say tanks with the button in the right upper window
Cheers
Borghese
  • Country: sg
« Reply #15 on: July 14, 2018, 03:55:59 pm »
OK I will do the other measurements after warmup,
can I ask why you think the signal doesnt look bad, I connected a 10 Mhz signal gen (sine wave, no modulation) to the input,
I would expect low (no) noise floor and a sharp peak at 10 Mhz instead I see all this mess ??!!
Offline borghese
  • Country: si
« Reply #16 on: July 14, 2018, 04:31:04 pm »
The image of post #12 look good.
Try with a signal generator a 10MHz O dBm. SA setting AT 20 .
Seem as you have some problem with lock as there is a lot of modulation
Cheers
Borghese
  • Country: sg
« Reply #17 on: July 14, 2018, 04:36:48 pm »
I took your recommended settings.
For the 10Mhz center where we do not see the noise floor the image is not stationary, the spikes are moving from right to left.
  • Country: sg
« Reply #18 on: July 14, 2018, 04:42:54 pm »
Hi, I am not sure I can follow.
My Signal Generator doesn't allow amplitude settings in dBm (or do I set this at the SA?).
Which setting do you want me to set to 20 ? The Amplitude ?
  • Country: sg
« Reply #19 on: July 14, 2018, 04:46:48 pm »
This is the screenshot
  • Country: sg
« Reply #20 on: July 14, 2018, 05:05:31 pm »
I found this regarding the LOCK OFF msg:
φ
LOCK OFF appears briefly during
the CAL FREQ routine, during instrument preset, or when the
frequency value is changed; this is normal and does not indicate a
problem.
Offline Bud
« Reply #21 on: July 14, 2018, 05:23:01 pm »
But it seems to stay and to me is an indication some  PLL has a problem. You should look in the schematic where they have a PLL.
Offline borghese
  • Country: si
« Reply #22 on: July 14, 2018, 05:44:00 pm »
In the post #12 the level of the 10 MHz signal seems OK; maybe you also have problems with the attenuator or bandwidth filters and PLL of course
« Last Edit: July 14, 2018, 06:35:09 pm by borghese »
Cheers
Borghese
  • Country: sg
« Reply #23 on: July 24, 2018, 04:33:15 pm »
I did a CONF TEST and the following attached errors were displayed,
would that point me to the right board to look for hardware errors ?
Offline borghese
  • Country: si
« Reply #24 on: July 24, 2018, 08:22:39 pm »
When was the last time you saw the analyzer work or is it a recent purchase?
Have you checked that all the voltage are between the limits?
It seems to me that you are not very familiar with the analyser
Cheers
Borghese