Hi Guys,
I have some progress to report... Yesterday I manufactured a 9w DSUB (M) to (F) cable with break out so that I could measure the ripple on the supplies to the modules and check out the serial bus from the controller, as this is used to set the control voltage on the 10MHz Reference Oscillator, which appears not to be working.
For the ripple I measured ~25mV P-P on the +12, -12 and +5V. Is this acceptable?
I confirmed that the clock and data to the Reference module were working (See attached). If I change the reference coarse and fine setting under the service menu I can see the bit pattern change. Apologies for the photo of the trace, the scope only has a GPIB interface and is no where near my PC.
Today I removed the Reference module from its metal case so that I could check that control voltage was actually being set.
It turns out the 12VDC supply +12V(F) to the VCO (U1) is missing, tracing it back I found that L2 was open circuit.
Measuring the supply pin on the oscillator it appears that it has developed an internal short (~34Ohms), which would have blown L2 on application of power.
This explains why the 10Mhz reference is missing and hence all the derived clocks. It also goes some way to explain the Self-Test failures, but I can only confirm this once I have located a replacement oscillator (P/N CTS KNIGHTS 970-4302-1).
Small steps....