HP8921A service monitor high stability reference noisy(fixed)
« on: September 12, 2015, 09:50:28 pm »
I recently traded some gear for an HP8921A service monitor. It had a problem which I narrowed down to the high stability reference board. I ordered a replacement board and that fixed the problem. The 8921A now works great. However that still leaves me wanting to fix the original board. HP has never released schematics
for the 8920/8921 series of service monitors which is a shame. They have released board level repair information only.
Enough information was provided to power up the board on the bench. It has a TCXO running at 10 MHz with a 1 GHz VCO that is phase locked and also divided by two to output a phase locked 500 MHz. The main problem is that the 1 GHz output is noisy, the 500 MHz is also less then perfect but doesn't look as bad after going through its divide by two circuitry. I have gone through the amplifier stages and everything seems to look good. It outputs the proper levels. The power supply to the oscillator itself looks good as does the control voltage. I can see it changing to adjust frequency and it looks pretty clean with no noise on it. The main VCO transistor is an MRF581. I have output scans from both the good and bad modules - you can see the noise on the bad one and a slight variation in its output frequency. The various traces are min/max/average and instant.

I have also included a pic of the board. It is pretty easy to see the flow of RF through the left hand side of the board starting at the top.
I have checked all resistors, power supplies etc, all look fine. If anyone has any thoughts let me know. I already swapped the MRF581 as I happened to have one and it made no difference.
« Last Edit: September 13, 2015, 06:34:17 am by TheSteve »
VE7FM
 

Offline TheSteveTopic starter

  • Supporter
  • ****
  • Posts: 3783
  • Country: ca
  • Living the Dream
Re: HP8921A service monitor high stability reference noisy
« Reply #1 on: September 13, 2015, 05:03:54 am »
Just a quick update.

I identified the bad part. It is U21, an MMIC amplifier. Marked A06 it is an Avago MSA-0686. It still seems to amplify but its input appears to no longer be 50 ohms which causes the VCO to be unstable.
« Last Edit: September 19, 2015, 04:17:49 am by TheSteve »
VE7FM
 

Offline py5lf

  • Newbie
  • Posts: 2
  • Country: br
Re: HP8921A service monitor high stability reference noisy(fixed)
« Reply #2 on: November 20, 2022, 07:01:32 pm »
Dear Steve
Do you have the schematic for HP8921A ? Mine are showing an alarm RF INPUT OVERPOWER and I want to fix it.
Thank you in advance
 

Offline Wallace Gasiewicz

  • Super Contributor
  • ***
  • Posts: 1362
  • Country: us
Re: HP8921A service monitor high stability reference noisy(fixed)
« Reply #3 on: November 20, 2022, 11:48:53 pm »
This group has a CLIP with lots of info
Here is the files address

https://groups.io/g/HP8924-Family-Spectrum-Analyzers/files/HP-8920A%20and%20HP8920B%20related/8920%20CLIP.pdf

I don't know if you have to join to read this, but its free and easy.
 

Offline Wallace Gasiewicz

  • Super Contributor
  • ***
  • Posts: 1362
  • Country: us
Re: HP8921A service monitor high stability reference noisy(fixed)
« Reply #4 on: November 20, 2022, 11:53:18 pm »
Steve:
I had a problem with another MMIC in a HP 8713 VNA. Just stuck in another MMIC that I had with the correct Freq specs and all is good.
I think it was the same MMIC (A06), in a different functional location,of course.
Wally
 