Hi
a few days ago, when I was using one of my 34401A, it started to show error 745 since powering up.
User Guide says that this error means: Cal checksum failed, FRES corrections.
Interesting is If I let the instrument warm up, after about 5 minutes when I shut it down and then I switch it back on, the error disappears. But if I cool down the EEPROM chip (U505) on PCB, this error comes back. So I took soldering pen (maybe cold joints), but no changes.
So do you think the EEPROM is faulty or (and I hope for this) it starts to "forget" ?
Last calibration was performed in Dec 1999 and I think that this cool down/warm up thing only changes Vgs of a transistor inside the EEPROM that carries the CAL informations (so sometimes the information is 1 and sometimes is 0), so it could be repaired only by rewriting -> performing new calibration. What do you think ?