Title: Hung Chang (Protek) 3502c oscilloscope
Post by: Mikeva on October 24, 2015, 07:35:06 pm
Hi guys,
I recently bought my first oscilloscope used and although all the functions work ok,I'm having problem with the vertical gain on the first channel.It can't measure the voltage right.
i connect the probe into the 0.5vcal point and it measures only 0.3Vp-p instead of 0.5Vp-p.The same happens if i try to measure dc voltage from my power supply.For 10 vdc it reads about 7.5-8v.The channel B  works ok instead.
I found the schematic and tried to calibrate it but the trimmer which is responsible for this is almost fully turned clockwise giving the measurements i described.
I checked all the voltages(except the 1.9kv :)) and all seems within the specs.
I didn't found any obvious fault on the components measurements.
There is an 15 pin hybrid ic marked AX-1015 (2 per channel) that maybe is the cause for the fault but i cannot find any information about it.Does anyone knows what ic is that?(pics)
Or any advice on how to proceed?
Any help will be greatly appreciated.
Thanks
 
Title: Re: Hung Chang (Protek) 3502c oscilloscope
Post by: siggi on October 25, 2015, 01:24:28 am
Input the same signal on both channels and trace it through, figure out where they diverge. Hopefully this happens in the discrete so before the hybrid.


Sent from my iPad using Tapatalk
Title: Re: Hung Chang (Protek) 3502c oscilloscope
Post by: Roeland_R on October 25, 2015, 03:17:45 pm
Does the cal potentiometer knob work correct?
Title: Re: Hung Chang (Protek) 3502c oscilloscope
Post by: Mikeva on October 25, 2015, 03:34:16 pm
yes it works alright .It has the same gain as the B channel.