Author Topic: Solved. 05-30-2020 Hameg Model HM203-6 Oscilloscope that is not working  (Read 1486 times)

0 Members and 1 Guest are viewing this topic.

Offline kastell123Topic starter

  • Contributor
  • Posts: 19
  • Country: es
  • Ramonesq
I have a Hameg Model HM203-6 Oscilloscope that is not working
Deformed (not straight) amplifier horizontal line (Y) is seen this in oscilloscope mode and in equal components mode, (not line) correct, I upload images to see.
Please help.
Thank you.
« Last Edit: May 30, 2020, 03:54:25 pm by kastell123 »
 

Offline kastell123Topic starter

  • Contributor
  • Posts: 19
  • Country: es
  • Ramonesq
Re: I have a Hameg Model HM203-6 Oscilloscope that is not working
« Reply #1 on: May 25, 2020, 07:42:08 am »
I have a Hameg Model HM203-6 Oscilloscope that is not working
Deformed (not straight) amplifier horizontal line (Y) is seen this in oscilloscope mode and in equal components mode, (not line) correct, I upload images to see.
Please help.
Thank you.
« Last Edit: May 25, 2020, 07:48:58 am by kastell123 »
 

Offline shakalnokturn

  • Super Contributor
  • ***
  • Posts: 2124
  • Country: fr
Re: I have a Hameg Model HM203-6 Oscilloscope that is not working
« Reply #2 on: May 25, 2020, 10:07:22 am »
Start by checking power supply voltages and ripple, this looks like interaction between X and Y (X deviation influencing Y) that should not normally happen.

Is it dynamic only or static and dynamic? If you set spot to center position in XY mode then shift it to left or right edge with position control or external DC input does it also move up or down?

Is there interaction the other way round? Does modulating the Y axis in XY mode also cause X axis movement?

Considering your sweep amplitude looks reduced I'd definitely start with checking power.
 

Offline kastell123Topic starter

  • Contributor
  • Posts: 19
  • Country: es
  • Ramonesq
Re: I have a Hameg Model HM203-6 Oscilloscope that is not working
« Reply #3 on: May 25, 2020, 05:55:19 pm »
Thanks for the answer, tomorrow 05-26-2020 I confirm your questions.
I think that the point, axes (Y) (X) if I can move them?
Just add for your information, that this breakdown suddenly occurred, one day I left it working fine (OK), and another day when I went to use it, was the breakdown present?
 

Offline kastell123Topic starter

  • Contributor
  • Posts: 19
  • Country: es
  • Ramonesq
Re: I have a Hameg Model HM203-6 Oscilloscope that is not working
« Reply #4 on: May 25, 2020, 06:19:29 pm »
Thanks for the answer, tomorrow 05/26/2020 I confirm your questions.
I think, by pressing the button that cancels deflection the point, axes (Y) (X) if I can move them
I can tell you for your information, that this fault, happened suddenly, one day I left it working fine (OK), and another day when I went to use it, it appeared, with the fault.
Sorry for my expressions, it's for the google translation
I understand something of the English language, but a little.
 

Offline kastell123Topic starter

  • Contributor
  • Posts: 19
  • Country: es
  • Ramonesq
Re: I have a Hameg Model HM203-6 Oscilloscope that is not working
« Reply #5 on: May 26, 2020, 08:46:30 pm »

    Hello
I have uploaded a video to Youtube for better fault compression.
Please open the link below.
Thank you.



 

Offline shakalnokturn

  • Super Contributor
  • ***
  • Posts: 2124
  • Country: fr
Re: I have a Hameg Model HM203-6 Oscilloscope that is not working
« Reply #6 on: May 28, 2020, 06:36:11 pm »
Have you checked all power supply voltages?
 

Offline kastell123Topic starter

  • Contributor
  • Posts: 19
  • Country: es
  • Ramonesq
Re: I have a Hameg Model HM203-6 Oscilloscope that is not working
« Reply #7 on: May 29, 2020, 07:50:05 am »
   
                    If all voltages ok except the voltage (-12) which are (-16.84 v)
 

Offline firewalker

  • Super Contributor
  • ***
  • Posts: 2450
  • Country: gr
Re: I have a Hameg Model HM203-6 Oscilloscope that is not working
« Reply #8 on: May 29, 2020, 08:08:22 am »
I had a similar problem with my HM-604. It was a faulty bridge rectifier on the power supply. AC was passing through the bridge and was creating issues like yours.

Alexander.
Become a realist, stay a dreamer.

 

Offline kastell123Topic starter

  • Contributor
  • Posts: 19
  • Country: es
  • Ramonesq
Re:Hameg Model HM203-6 Oscilloscope that is not working
« Reply #9 on: May 29, 2020, 08:18:53 am »
 All voltages ok except voltage (-12) which are (-16.84v)
Power source, to bien except (-16.84) in brand scheme (-12 v)
Sorry there are some, voltages on circuits with errors.
In the final horizontal amplifier (X Final Amplifier) in the channel that attacks the deviation plate (D1) there is + 99 v, and in (D2) + 131, obviously there is an error, the two voltages must be almost the same with some difference of 2… 5 volts No (32 volts) it seems that there is no consumption on the board (D2) or I have some pre-amp before the final amplifier, with defect (some Bf199) the Bf458 and BF472 checked and well, I will continue investigating 1 (-16.84) and check BF199.
Why doesn't the point appear when I press the (X-Y) button… ??
« Last Edit: May 30, 2020, 07:12:07 am by kastell123 »
 

Offline kastell123Topic starter

  • Contributor
  • Posts: 19
  • Country: es
  • Ramonesq
Re: I have a Hameg model HM203-6 Oscilloscope that is not working
« Reply #10 on: May 29, 2020, 08:25:43 am »
     
              Thanks, Alexander
I'll check that bridge, rectifier.
« Last Edit: May 30, 2020, 07:14:42 am by kastell123 »
 

Offline kastell123Topic starter

  • Contributor
  • Posts: 19
  • Country: es
  • Ramonesq
Solved. 05-30-2020 Hameg Model HM203-6 Oscilloscope that is not wor
« Reply #11 on: May 30, 2020, 03:44:46 pm »
 
     Solved. 05-30-2020
The whole problem was because of this cross capacitor(c714 0.1UF (100K 63V) (short, total) now there is (- 12 volts)
Before (-16.84 volts) this was the whole problem, just this shorted capacitor, nothing else.
Attached diagram for clarity of the affected component.
This afternoon I will upload 2 videos, to see the solution.
« Last Edit: May 30, 2020, 03:51:36 pm by kastell123 »
 

Offline kastell123Topic starter

  • Contributor
  • Posts: 19
  • Country: es
  • Ramonesq
Elace condensador culpable de averia.
   
Solucionado 30-05-2020


 

Offline jdragoset

  • Regular Contributor
  • *
  • Posts: 242
  • Country: us
 :-+
Great "sticking to it" work.
Got a way-to-cheap Tek 067-0532-00 Constant Amplitude Calibration Fixture ( with GR 1363 unit oscillator to 500 megahertz)
Q707 (SCR) was missing but could momentarily short A to K and start the unit.
Saw above 100 mhz sine wave but shortly vanished with smoke/heat around the 40 uf, 500 volt dual-section Mallory electrolytic. Got a dual section, 50-uf, 500 volt replacement for under $20
Same size and the original poly sleeve actually fit.
Adding an NTE 5410 SCR would still not start but found Q 711 (SCR gate quench) shorted.
Replaced with a 2N5961 and all works.
Off subject but never give up!
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf