No worries. as long as your DMM and scope didn' did this:

that could be fixed.
One good rule is to remember the original pot settings before changing them or take a picture (which might help for simple pots, not for 10-turn trimmers). If it's too late: You might be lucky, you have a complete manual of your scope. The bad thing is, depending how many trimmers you touched this can take some time and good understanding of the meaning of the individual knobs.
You should try to understand the schematics a bit. Moving the POS doesn't change anything there as the POS poti is located later in the signal path. So Here the poti R543 and 547 might change this. But this looks like only the trigger supposed to be affected. That should have no impact to the traces on the screen. We should save this for later (but you should keep this in mind).
Can you please post what you measure at TP4 - TP5 (see picture)? As you see we're down to few mV/div offset. Unfortunately the 7.8mV/div don't say to GND or to the 'other side', so please measure the voltage between TP4 and TP5 as this might be double the value thus easier to measure. Because you wrote earlier that the values on the anodes V521/522 showed some offset (13mV) this might match with the offset you see. You 'only' need to figure out the first place of this mismatch and check what could be the cause there.