Author Topic: Identify component on TDK CCFL inverter  (Read 1212 times)

0 Members and 1 Guest are viewing this topic.

Online bdunham7Topic starter

  • Super Contributor
  • ***
  • Posts: 7825
  • Country: us
Identify component on TDK CCFL inverter
« on: August 14, 2020, 06:20:19 pm »
I have a TDK CXA-L0605-VJL CCFL inverter board from a R&S CMU200 that I'm repairing.  I've already ordered a new/used/refurb board but it won't be here for a while and if I can fix this one, I'd like to.

The problem appears to be the component marked '20' in the photo.  Bypassing it allows the board to work and it appears to function OK.  I'm assuming this is an SMT fuse or PTC device, but I'm not sure which. I was unable to locate that code for any device that made sense.  If I repair the board, I want to use the correct device, not the small wire strand I have bodged across it right now.  The device draws about 950mA in operation, so I'm assuming this is a 2A or so device.  It is 1210 size.  Does anyone know exactly what it is?
« Last Edit: August 14, 2020, 06:24:10 pm by bdunham7 »
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Online bdunham7Topic starter

  • Super Contributor
  • ***
  • Posts: 7825
  • Country: us
Re: Identify component on TDK CCFL inverter
« Reply #1 on: August 14, 2020, 07:18:38 pm »
OK, I found it.  It is a KOA CCP2E20.  This is a 1210 fast-blow fuse with a 2A fusing current (blows in 1s) and an 800mA continuous current rating.  IMO, that's problematic for a unit that draws 950mA in normal operation, so I will put in a 1.25A replacement with a 2.5A fusing current.  I'll leave the post up since a lot of people here have CMU200s.  If you have a dark screen, order a fuse before you go to all the trouble of tearing it apart because I'm betting I'm not the only one.
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline shakalnokturn

  • Super Contributor
  • ***
  • Posts: 2123
  • Country: fr
Re: Identify component on TDK CCFL inverter
« Reply #2 on: August 14, 2020, 10:33:14 pm »
Is it possible that the 950mA arn't normal operation?

Could be caused by transformer insulation breakdown, leaky ceramic cap on secondary, PCB or wire insulation breakdown.
I've seen breakdown of heatshrink sleeving over the ends of the CCFL's before in other equipment.
 

Online bdunham7Topic starter

  • Super Contributor
  • ***
  • Posts: 7825
  • Country: us
Re: Identify component on TDK CCFL inverter
« Reply #3 on: August 14, 2020, 10:54:43 pm »
Is it possible that the 950mA arn't normal operation?

Could be caused by transformer insulation breakdown, leaky ceramic cap on secondary, PCB or wire insulation breakdown.
I've seen breakdown of heatshrink sleeving over the ends of the CCFL's before in other equipment.

The inverter board is an off-the-shelf unit used in a specific application.  I have two different data sheets for  the unit--one says typical 700mA, max 800mA but the second says typical 800mA and max 1000mA.  I have no information about the normal current draw for the specific application (CCFL type) so it is possible that the unit is just run near its maximum in this application.  It seems to work fine and I'm giving it a long burn-in test.  However, even if you go by the first data sheet, running a fuse at 100% of its continuous current rating is just pushing the limits a bit too much, IMO, unless you really need a low fusing limit.  In this case, I don't think there's any need for an extremely low current fuse as the circuit is driven off of essentially a PC power supply.
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline KenPaisley

  • Newbie
  • Posts: 3
  • Country: us
Re: Identify component on TDK CCFL inverter
« Reply #4 on: October 23, 2020, 02:01:58 am »
CFL tube and what appears to be genuine TDK inverter is available on Amazon. My display is very bright now. Don't bother with the inverters on eBay. They work, but display is dim. $50 for the inverter and $22 for the lamp.
 

Online bdunham7Topic starter

  • Super Contributor
  • ***
  • Posts: 7825
  • Country: us
Re: Identify component on TDK CCFL inverter
« Reply #5 on: October 23, 2020, 02:13:57 am »
Do you have a link for the CCFL tube?
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline KenPaisley

  • Newbie
  • Posts: 3
  • Country: us
Re: Identify component on TDK CCFL inverter
« Reply #6 on: October 23, 2020, 02:18:58 am »
https://www.amazon.com/gp/product/B07L4Z6N3X/ref=ppx_yo_dt_b_asin_title_o00_s00?ie=UTF8&psc=1
My other cmu200 I had to recap the motherboard and the made the screen brighter than it was, but I ordered another cfl tube. Hope that link works.
 

Offline KenPaisley

  • Newbie
  • Posts: 3
  • Country: us
Re: Identify component on TDK CCFL inverter
« Reply #7 on: October 23, 2020, 02:24:58 am »
« Last Edit: October 23, 2020, 02:27:48 am by KenPaisley »
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf